DLCO overestimated from an apparent zero offset error

I’ve had some concerns for a while now about how the CO and CH4 concentrations are being calculated from the DLCO analyzer calibration zero offsets and gains on our test systems. For this reason I’ve been looking carefully at all of the raw data from our DLCO tests and today I came across an oddball test result. There are several reason why this is probably not the best example for this particular problem that I could come up with but it illustrates an important point and it’s in front of me so I’ll go with it.

In order to use the output from a gas analyzer you need to know the zero offset and the gain of the signal. Presumably the analyzer remains stable enough between the time it was calibrated and the time it is used for the zero offset and gain to be meaningful. When looking at the calibration data I’ve noticed that some of our test systems show relatively large changes in zero offset from day to day. These changes are still within the operating limits of the analyzer so no red flags have gone up over this. The test systems and analyzers are turned off over night so in order to see if the analyzers go through these kind of changes during a normal day I once did a series of calibrations each separated by five or ten minutes on one of the more suspect testing systems. What I saw was that although there were small changes from calibration to calibration, I didn’t see anywhere near the changes I’ve seen from day to day which at least implied that the analyzer remained reasonably stable during a given day.

Today a patient’s report came across my desk and as usual I took a look at the raw test results. What I saw was that two out of three of the DLCO tests had been performed with the correct inspired volume but that the one with a much lower inspired volume had a much larger VA and DLCO when compared to the other results. This got me scratching my head since the patient has severe COPD and that usually means that a lower inspired volume leads to a lower DLCO and VA. When I noticed the analyzer signals during the breath-hold period that’s when I could see right away why the results had been overestimated.

Here’s the test with the low inspired volume:

DL_Gain_Error_01_Graph_Datasheet_Cropped_2c 

Here’s a test from the same patient with the proper inspired volume:

DL_Gain_Error_01_Graph_Datasheet_Cropped_2b 

In our test systems the DLCO analyzer continues to sample the inspired gas during the breath-holding period. On the graphs this is normalized using the calibration zero offset and gain to 100% rather than to a specific concentration (it doesn’t matter what values you use because it’s the ratios of concentrations that matters not the actual concentrations). In the oddball test the inspired gases show up as being around 85% not 100%. You will also notice that in this test the exhaled CO and CH4 concentrations are significantly lower than in the other tests.

DL Offset table

My take on this (and of course the manufacturer’s technical staff may well disagree) was that there was an abrupt decrease in the DLCO gas analyzer’s zero offset. When the zero offset decreases, even if the gain remains unchanged, the analyzer’s signal output will be reduced. This is what I think caused the inspired gas concentrations to be reduced.

Although the DLCO gas analyzer goes through a pre-test check (not a calibration), either the change in zero offset was still within specifications or the zero offset changed after the pre-test check. If the analyzer’s signal was reduced then the exhaled CH4 and CO concentrations will also be reduced. Since VA is calculated from the inspired volume and the change in CH4, despite the fact that the inspired volume was low the calculated VA is elevated. DLCO is in turn calculated from the VA and the change in CO and since the VA was elevated and the exhaled CO was reduced the DLCO is going to be higher.

The oddball test was the second of three tests. The first and the third test both had the proper inspired volume and the inspired gases were at 100% so whatever the problem was, it came and went quickly. It’s possible there was a voltage surge or dropout although I would have expected the power supply for the analyzer to handle these things. This is one of the reasons that makes this test a poor example. What it does illustrate however, is that (once again) the test with the highest results is not always the test that should be reported. It also shows the need to remain vigilant about even small details in test results.

I think that inspired volume is the most important quality indicator of DLCO tests. There is more than sufficient reason to be suspicious when a DLCO test with a low inspired volume has a higher result than a test with the proper inspired volume. Although I also tend to think that a DLCO test with a higher VA is probably more accurate than a test with a low VA in this case I think the elevated VA was due to an analyzer error and the clue to that error was the low inspired gas concentration.

This is likely a moderately unusual error at least in terms of its magnitude. It remains unclear to me just how common or uncommon this kind of problem actually is. I suspect that on a much smaller scale it is probably a common occurrence since that’s just the nature of analog electronics. It’s taken me some time but I’ve learned from our equipment manufacturer that the software for our test system uses the zero offset and gain from the last “real” calibration to calculate exhaled CH4 and CO. Even though the DLCO analyzer goes through a pre-test check, the results of check are compared to the normal operating range of the analyzer and not to the last calibration. Since the results from the pre-test analyzer check are not saved or stored in any way this means that (at least presently) it’s not possible to determine what changes in zero offset and gain routinely occur during the course of a given day.

[Warning, rant ahead!]

In a more general sense I am concerned that the details of how our pulmonary function test equipment actually gets from physical measurement to numerical results has become, if not exactly hidden, at least difficult to get at. I don’t necessarily blame the equipment manufacturers because if more people in our field asked these kinds of questions they would likely be more forthcoming with answers. Having said that I don’t think it is realized just how much of the accuracy we take for granted in the test systems we use every day is based on proprietary, and therefore opaque, hardware and software processes. I wince every time I read a research paper and see that critical results came from “test system model 123 of manufacturer X” and it is apparent the accuracy of the equipment was never questioned or verified. I would really would like to see a lot more skepticism on the part of researchers, technicians and medical directors as well as more openness from the equipment manufacturers.

Creative Commons License
PFT Blog by Richard Johnston is licensed under a Creative Commons Attribution-NonCommercial 4.0 International License.

4 thoughts on “DLCO overestimated from an apparent zero offset error

  1. Hi,
    Im chandra,M.Tech from IITH ,present we r working on spirometer,we r extract the 7 paremeters from normal spirometer…now we r going to DLCO measurment…so kindly help to move further steps….like which type of sensors want to use ,wat confg,and how much sensitivity…

  2. Your test system looks like to be a fast response analyzer possibly SensorMedics or Collins. What I do not understand is why gas concentrations are going high before IVC, as upon onset of IVC demand valve should open and patient inspires gas concentration and sample line takes gas sample from the mouth to the analyzer! Besides, if the two graphs are of the same patient, CH4 concentration during collection volume (sample volume) is
    different. In the upper graph it is strsight line and in
    the second graph with a slope, which is indication of patient with distribution problem!
    This is in my opinion the source of differences in measurement results and not necessary offset changes of the anslyzer. As you probably know there is a method of Fast Space back extrapolation to accurately measure RV and TLC from DLCO test.

    • Saeed –

      Yes, the graphs are from a system with a fast gas analyzer and they are both from the same patient. The graph goes high because the balloon valve in the patient manifold opened, but the patient hadn’t inspired yet. About a half a year after I wrote this blog I had an opportunity to sit down with the head software engineer for the company. It turns out to be a combined mechanical and software error. The software looks at the analyzer signals at the beginning of the test and averages the CH4 and CO signals under the assumption that they are sampling the 100% DLCO test gas mixture at that point, and uses this to set the offset (not the gain) of the 100% signal. The demand valve was not opening properly and allowed a puff of room air to get into the system and the signal from this was averaged as well. The software did not change the gain, which comes from a prior calibration, just where it thought the 100% signal was. This error then essentially causes the zero to shift downwards, but it is a software error that does this, not the analyzer. Since the software still used the “real” zero from a prior calibration this means that the CH4 and CO results were both underestimated which caused an erroneously high VA and DLCO to be calculated.

      What we see on the computer screens is not always an exact representation of what the gas analyzer and flow sensor are seeing. This is always mediated by software and in this case, it was affected by, if not an error, then at least a mistaken assumption by the software (and the programmer). The company claims this error has been corrected and that we will get the fix in our next software update.

      FYI, I have seen a variety of schemes and correction factors to calculate “real” TLC and RV from a DLCO test. When averaged over a large group this appears to work, but when you look at individual results there is actually a very wide scatter in the results. VA is always underestimated (when compared to TLC) in the presence of airway obstruction and ventilation inhomogeneities and the actual location and mix of these factors varies from one individual to another even when they have the same FEV1 and FEV1/FVC ratio. For these reasons I remain skeptical of any method used to extrapolate TLC and RV from the DLCO maneuver and do not feel that they can substitute for actual lung volume measurements.

      Regards,

      – Richard